Nuclear Pioneer MCA RIID gamma spectrometer

Messaggio da Matrix » 05/11/2021, 23:15

Ciao amici,
I would like to represent my development of gamma spectrometer:
Amateur based RIID gamma spectrometer for spectrum acquisition, isotope identification and activity calculation.
Supports NaI(Tl) or CsI(Tl) crystals 63x63, 51x51, 40x40, 25x25, 10x40

Isotope Library Include: Cs-137, Ra-226, Th-232, Bi-214, Ba-133, Am-241, Lu-177, K-40, Co-60, U-235
Activity calculator for: Cs-137, Ra-226, Ba-133, Am-241
  • 12-bit 4096 channels ADC
    TFT LCD 3.5” 480x320 with adjustable backlight level
    Rotary Encoder for controls
    600V-1000V adjustable high voltage for photomultiplier probe positive polarity
    400V or 500V high voltage for an internal Geiger Counter
    Adjustable gain & pulse shape pre-amplifier for an external photomultiplier probe
    Gamma Spectrometer Mode (tested up to 10.000 CPS)
    Scintillation Counter Mode with PMT and GM Sensor
    Easy Channel to Energy calibration with selectable scheme and channels pick
    Real time background subtract mode for gamma spectrometer
    Real time gaussian peak detection
    Real time activity calculation (beta since 4.07 firmware)
    Linear or Logarithmic spectrum viewer
    Energy compensated dose rate
    0.30 uSv/h doserate alert and 2-sigma CPS alert in scint.mode
    CPS standard deviation calibration for scint.mode
    Mute / Alert / Clicking in scint.mode
    Spectrum LCD zoom 512-4096 channels
    Micro SD card for data storage
    RTC (require CR2032 battery for date and time keeping)
    Internal temperature sensor
    Bluetooth BLE module (optional)
    12Mbit USB port for programming, calibration or data transfer
    Easy firmware update via USB and friendly bootloader
    Power supply: 3.7V LiPo battery, battery current consumption 250mA-350mA, LiPo included
    3.7V LiPo battery USB charger, software low battery alert
Immagine

[External Link Removed for Guests]

[External Link Removed for Guests]


[External Link Removed for Guests]

Avatar utente
max56fe
Messaggi: 292
Iscritto il: 13/03/2016, 16:53
Località: Ferrara

Re: Nuclear Pioneer MCA RIID gamma spectrometer

Messaggio da max56fe » 07/11/2021, 8:54

and luckily what do you say amateur !!! you did a huge job really congratulations, only thing I do not understand why HT limited to 600-100 and not 400-1500 for the rest is really baby, congratulations again, I hope that one day it is also on sale and I book already from now , ciao and good luck


Tutto a questo mondo ha una fine...................... solo la salciccia ne ha 2 !

Avatar utente
cipndale
Messaggi: 503
Iscritto il: 16/01/2016, 18:41
Località: Napoli

Re: Nuclear Pioneer MCA RIID gamma spectrometer

Messaggio da cipndale » 07/11/2021, 9:55

[Local Link Removed for Guests] ha scritto: [Local Link Removed for Guests]
07/11/2021, 8:54
and luckily what do you say amateur !!! you did a huge job really congratulations, only thing I do not understand why HT limited to 600-100 and not 400-1500 for the rest is really baby, congratulations again, I hope that one day it is also on sale and I book already from now , ciao and good luck
Nice effort. For max56 it is already on [External Link Removed for Guests].
HV is probably limited due to materials used and the distance between the BNC leads on the PCB board. That distance would certainly be gapped at 1500 V. Also BNC is not rated above a certain voltage.



Avatar utente
cipndale
Messaggi: 503
Iscritto il: 16/01/2016, 18:41
Località: Napoli

Re: Nuclear Pioneer MCA RIID gamma spectrometer

Messaggio da cipndale » 07/11/2021, 10:08

[Local Link Removed for Guests] ha scritto: [Local Link Removed for Guests]
05/11/2021, 23:15
Isotope Library Include: Cs-137, Ra-226, Th-232, Bi-214, Ba-133, Am-241, Lu-177, K-40, Co-60, U-235
Activity calculator for: Cs-137, Ra-226, Ba-133, Am-241
Hi Matrix,
Can you detail a realistic case where you can approach the U-235 identification with this system?
What about the NORM detection, meaning above 1.5 MeV; can the system handle well that range?
Also, do you have some linearity correction implemented?
Regards!



Avatar utente
Matrix
Messaggi: 21
Iscritto il: 10/08/2018, 14:30
Località: Jerusalem
Contatta:

Re: Nuclear Pioneer MCA RIID gamma spectrometer

Messaggio da Matrix » 07/11/2021, 10:44

[Local Link Removed for Guests] ha scritto: [Local Link Removed for Guests]
07/11/2021, 8:54
and luckily what do you say amateur !!! you did a huge job really congratulations, only thing I do not understand why HT limited to 600-100 and not 400-1500 for the rest is really baby, congratulations again, I hope that one day it is also on sale and I book already from now , ciao and good luck
Thank you. The HV above 1kV can be a problem for regular BNC jack adapter that I'm using. Direct soldering of a cable to the pcb can solve this limitation for me, but will a problem for someone who use the kit.


[External Link Removed for Guests]

Avatar utente
Matrix
Messaggi: 21
Iscritto il: 10/08/2018, 14:30
Località: Jerusalem
Contatta:

Re: Nuclear Pioneer MCA RIID gamma spectrometer

Messaggio da Matrix » 07/11/2021, 11:03

Hi Matrix,
Can you detail a realistic case where you can approach the U-235 identification with this system?
What about the NORM detection, meaning above 1.5 MeV; can the system handle well that range?
Also, do you have some linearity correction implemented?
Regards!
1. you can view the screenshots of the spectrum with detected isotopes. If you are talking about natural ore samples where mostly U-238 is presented, the device cannot detect U-235 in the mix. But if you have a fiestaware or uranium glass it should trigger U-235 detection well.
IMG_5033.JPG
2. the system can handle the high energy range, depend on the gain calibration of the hardware. The total gain is set with HV and pulse amplifier.
3. energy scale is calibrated with 3 points polynomial that compensate some non-linearity of the probe. However the linearity mostly depend on the PMT specifications and hardwrae gain, you need to check that system has accpetable linearity.
Here I'm attaching exported CSV opened with BecqMoni. As you can see the a coefficient of the polynom is very small that say linearity is great.
eu-152-5151.jpg
Non hai i permessi necessari per visualizzare i file allegati in questo messaggio.


[External Link Removed for Guests]

Avatar utente
max56fe
Messaggi: 292
Iscritto il: 13/03/2016, 16:53
Località: Ferrara

Re: Nuclear Pioneer MCA RIID gamma spectrometer

Messaggio da max56fe » 07/11/2021, 17:25

hi cipndale, thanks for the connection to the commercial, as you well know both BNC RP-SHV reach 5Kv, I use branded BNCs and non-Chinese ones up to 2 Kv without problems, important is the RG58 / 59 cable that is use, after many tests I only use Messi & Paoloni double-shielded and decidedly soft cables, they are called ULTRAFLEX 7 and also have excellent connectors
[External Link Removed for Guests]


Tutto a questo mondo ha una fine...................... solo la salciccia ne ha 2 !



Avatar utente
cipndale
Messaggi: 503
Iscritto il: 16/01/2016, 18:41
Località: Napoli

Re: Nuclear Pioneer MCA RIID gamma spectrometer

Messaggio da cipndale » 09/11/2021, 21:44

[Local Link Removed for Guests] ha scritto: [Local Link Removed for Guests]
07/11/2021, 11:03

3. energy scale is calibrated with 3 points polynomial that compensate some non-linearity of the probe. However the linearity mostly depend on the PMT specifications and hardwrae gain, you need to check that system has accpetable linearity.
Here I'm attaching exported CSV opened with BecqMoni. As you can see the a coefficient of the polynom is very small that say linearity is great.

eu-152-5151.jpg
Ok
Thank you for the reply. Just out of curiosity is that the NIST Eu152 standard?